In the above post we discussed the various background methods available in Probe for EPMA (traditional off-peak, Mean Atomic Number and Multi-Point Backgrounds). In this post we will start discussing various considerations for selecting one background type over another, given specific analytical situations. Let's start by listing some factors in our decisions:
A. Acquisition time (many analysis points vs. only a few analysis points)Although using the MAN background method saves about half the acquisition time per point, if one is only planning a few dozen analyses it might not be worth the time and effort to acquire an MAN curve. On the other hand, if one acquires all elements using MAN curves, there is no need to acquire wavescans to check for off-peak background interferences or absorption edges. However once one is acquiring more than a few dozen point analyses, the MAN background begins to save considerable acquisition time (and provides better precision than off-peak methods). Note also that because a standard can be acquired once and utilized both as a primary standard and an MAN standard,the amount of acquisition time required for the MAN curve can be only a few minutes.
B. Availability of pure standards (for MAN curve generation)However, if one is going to utilize the MAN background method, it really helps to have a number of pure standards covering a range of Z-bar (average atomic number) that includes the standards and the unknown samples under consideration. But usually the analyst can find a few pure oxides (or even pure elements or metals) that can be utilized for the MAN curves. Ideally one would like three or more standards for each analyzed element (that do not contain the element in question). However, alternatively one can utilize impure standards (and even standards containing the element!) and acquire off-peak backgrounds on the standards, and utilize MAN backgrounds for the unknown samples, by utilizing the "interpolated" MAN below) as described here:
https://probesoftware.com/smf/index.php?topic=987.0This method allows the user to acquire off-peak backgrounds on ones standards, and then use the *interpolated* off-peak intensity for constructing the MAN correction curve. This interpolated MAN curve can then be applied to unknown samples acquired using the MAN background method.
C. Complex or curved backgrounds or with many over-lapping peaks (e.g., REE analysis)Samples with very curved and/or complex backgrounds require additional considerations. The multi-point background (MPB) method, first described by Mike Jercinovic and Julien Allaz, allows the user to specify more than one off-peak background positions on each side of the emission line for a more accurate determination of the background intensity under an emission line. This is especially useful for highly curved backgrounds at low sin thetas, or when the sample is highly heterogeneous and/or there are no ideal places to measure the off-peak backgrounds.
The MPB method does require more time for acquisition of the additional backgrounds, but provides enormous flexibility in fitting the background subsequent to acquisition. Think of it like acquiring a high precision (but very sparse) wavescan, at the same time as the unknown acquisition! This background method allows one to skip tedious wavescans on multiple samples and yet provides a highly accurate background determination.
The MAN method can also be considered for such complex background situations, but this partly depends on the standards available for the MAN curve.
D. High accuracy trace element analysis (is there a suitable blank standard available)Regardless of whether one utilizes the off-peak, MAN or MPB method for background acquisition, the use of a blank sample ensures that one trace element accuracy is *equal *to the measurement precision as described here:
https://probesoftware.com/smf/index.php?topic=29.msg387#msg387A blank sample is of course a standard with a matrix that is roughly similar to the sample in question, but which contains a *zero* concentration of the element in question, and is acquired as an unknown sample, using the *same* acquisition method as the unknown samples, to which it will be applied. By a zero concentration we mean a concentration lower than the instrument is capable of detecting. So for an EPMA instrument, generally any concentration below 1 PPM can be considered a "blank" level. I still refer to Mike Jercinovic's admonition: "If you can't measure something, then see if you can measure nothing... because, if you can't measure nothing, then you can't measure anything."
Of course the blank correction in Probe for EPMA can also utilize standards with a known non-zero concentration of the element in question. But then we come back to problems with using trace element standards (with a non-zero concentration of the element). Specially we shouldn't be using trace element standards because we really don't know the trace concentration of an element. Is the concentration 100 PPM or is it 110 PPM, and more importantly, how do we know? Yes, we can use another method such as ICP-MS and get an average of the trace element, but then how well do we know that the trace element is homogeneous at the micro-scale? Of course the irony is that most techniques (such as ICP-MS) run a "blank" sample to determine zero anyway. So why don't we just do the same thing on the EPMA using a blank standard?
E. Beam sensitive materials (TDI vs. short count times)If time is of the essence because our sample is very beam sensitive, then the MAN method can be helpful to limit the beam exposure of the sample.
In fact even when using the MPB method Mike Jercinovic, Julien Allaz and Karsten Goemann will often utilize the Nth point background method, where the first point is a "sacrificial" point utilized simply to determine the background intensity. Which is then applied to each subsequent point assuming that the sample is relatively homogeneous. More details on the Nth point method is here:
https://probesoftware.com/smf/index.php?topic=806.msg8036#msg8036Writing this reminds me of the old ARL SEMQ EPMA instrument we had at Berkeley which had 4 fixed monochromators (Si, Fe, Ca, Al) and 4 tunable spectrometers, so using the MAN background method we could acquire 8 elements in 10 seconds. Which was very useful for measuring alkali rich glasses.
F. Average Z of standards and unknowns (high Z materials = high continuum intensities)Finally there is the issue of average atomic number or average Z or Z-bar. As we know from Kramer's Law, the continuum intensity is proportional to the average Z of the material. In silicates and oxides (and many mineral glasses) this usually means a range of Z-bar from 10 to 20. E.g., MgO, Al2O3 and SiO2 are about 10, TiO2 is around 16 and MnO and Fe2O3 are around 20. So if you have pure standards like these, it is very easy to acquire an MAN curves suitable for almost all oxides, silicates and glasses.
Using MAN backgrounds in such materials we usually see accuracies around 100 to 200 PPM or better. And of course with a blank standard/sample even better than that.
However, as we look at materials with higher Z-bars, the continuum intensity increases and our P/B tends to go down, meaning that the accuracy of the background correction becomes even more critical. While this is true for all background measurement methods, the advantage of the MAN method remains that one is directly measuring the continuum at the emission line position so no interpolation (or extrapolation) from off-peak measurements is necessary. Therefore there are no issues with off-peak interferences, peak tails or absorption edges.
On the other hand, the availability of pure high Z standards for calibrating the MAN curve becomes the dominant question for high Z samples. And while the MAN curve is not affected by off-peak interferences, it most certainly is affected by *on-peak* interferences. Luckily, whether these "interferences" observed in our MAN curves is due to an actual on-peak interference, or a (previously unspecified) amount of the element in the standard, we can simply apply the rule of thumb that "background is (by definition) generally the lowest intensity we can measure". So if we see a high intensity outlier in our MAN curves, we can simply reject that intensity and re-fit the MAN curve. Here are some examples of "outliers" in MAN curves:
https://probesoftware.com/smf/index.php?topic=4.msg499#msg499And here are some examples of a few of these high Z-bar MAN curves are seen here:
https://probesoftware.com/smf/index.php?topic=4.msg5136#msg5136And once again, remember that one can obtain accuracy in the MAN method equal to one's precision simply by apply a suitable blank standard/sample to the analyses. This becomes especially important the higher the Z-bar of ones samples.
Of course one can also utilize a "mixture" of background acquisition methods. For example, acquire the major elements using MAN backgrounds, and traces using a higher beam current and off-peak backgrounds. What are your own thoughts regarding the selection of background methods?